Risk-benefit analysis is the comparison of the risk of a situation to its related benefits.
The FDA provided a guidance about the benefit and risk factors to be considered for making decisions concerning product availability, compliance and enforcement.

Can you tell us one of those factors and explain why it is important ?

For example:
Assessment of the Benefits of Devices includes: The type of benefits, The magnitude of the benefits, and how long the benefits can be expected to last for the patient.

Hint:
https://www.fda.gov/downloads/medicaldevices/deviceregulationandguidance/guidancedocuments/ucm517504.pdf


 
Posted : 13/11/2017 10:03 am
(@amandaally1029)
Posts: 40
Trusted Member
 

one factor to consider is the assessment of the risks of the device. With this assessment you can determine the severity and number of harmful events associated with this device. Also, the probability of harmful events, duration, and the risk from false-positive or false-negative results for diagnostics. The reason why this factor is important is because during analysis, it can help determine whether the device risks are severe or moderate. IT can also help determine whether the device should even be available to people based on its risks. It's best to figure these things out ahead of time before it causes real harm, and by making this assessment it can help you determine if this device is acceptable or not.


 
Posted : 13/11/2017 10:54 am
(@mark-abdelshahed)
Posts: 80
Trusted Member
 

Another risk factor to be considered for making decisions concerning product availability, compliance and enforcement is factors for the assessment of medical device benefits. Benefit, as described by the potential benefit factors, may change over time. FDA may assess the extent of benefit of a device by considering factors like: type of benefits, magnitude of benefits, likelihood of patients experiencing one or more benefits, duration of effects, patients perspective on benefit, benefit factors for healthcare professionals or caregivers, and medical necessity. It is important to compare and contrast the benefits and risks identified based on the available information, in order to see if the benefits of this specific device is more than the risks of it or not.


 
Posted : 13/11/2017 2:36 pm
(@dag56)
Posts: 79
Trusted Member
 

To add to what was said above by both Mark and amandaally1029 in the above comments, another factor to consider would be the “Availability of alternative treatments or diagnostics”. This is a critical factor to consider because it would not only directly affect the need for your product, but the likelihood of the FDA signing off on the project. If other treatments are readily available to treat the same symptoms cheaper, then it would not be financial beneficial to design and produce the device in question. Similarly, if a device already exists that is able to treat the same symptoms with lower possible health risks, the FDA will not sign off on the proposed new project idea. Ideally, the proposed device would fill a major need in the market with little to no other comparable competition in both the financial and benefit sectors. Lastly, it is important to note that although a product may be deemed to have potentially dangerous side effects towards the patient, the FDA may still sign off on the device if there are no other options available on the market. This goes by the hope that the “the probable benefits outweigh the probable risks”.


 
Posted : 14/11/2017 7:51 am
(@smitshah)
Posts: 75
Trusted Member
 

Factors to be considered for benefits vs Risk analysis by analysis.

Uncertainty:is an important factor, since at any point in the total product life cycle, there is never
100% certainty regarding the safety, effectiveness, or quality of a device. However, the degree of
certainty of the benefits and risks of a device is a factor FDA considers when making benefit-risk
assessments.

Mitigation:are actions taken by the manufacturer, by FDA, or by other stakeholders to recover
benefit or to limit harm. Mitigations could address, among other considerations, as applicable:
clinical practice; use errors; unmet medical needs; the use environment; user population; user
skill level; clinical understanding in assessing risk; current expectations in clinical use; any
changes in medical practice etc.

Detectability refers to whether nonconformity could be identified, either by the manufacturer
or by the user. A nonconformity which can be identified prior to use of the device may harm
fewer patients than a nonconformity which is not identified prior to use.

Failure mode is the specific method or type of failure. The failure mode may be used to identify
the cause of the nonconformance, including whether the nonconformance is related to
manufacturing, design, use conditions, or environment.

Scope of the device issue should be evaluated to assess whether the risks identified are potentially
inherent to similar devices of this type

Patient impact is the impact on the health and quality of life of patients if a particular compliance
or enforcement action is, or is not, taken or if the device relevant to the nonconformity or
regulatory non-compliance is not available.

Firm compliance history may include the manufacturer’s regulatory history and initiative in
identifying and correcting issues, the repetitiveness of such issues, and the manufacturer’s
communication with FDA.


 
Posted : 15/11/2017 9:28 am
(@thuytienlecao)
Posts: 72
Trusted Member
 

I would like elaborate @merzkrashed's point about the FDA article "Factors to Consider When Making Benefit-Risk Determinations in Medical Device Premarket Approval and De Novo Classifications".
When determining risk, the FDA assessing the severity, type, frequency, and length of harmful events. In addition, the FDA will also evaluate other factors such as patient-centric assessments and patient-reported outcomes, characterization of the disease, alternative treatments like @dag56 has mentioned and risk mitigation strategies like @smitshah have mentioned. An important point is that the FDA also stated that "risk tolerance will vary among patients, and this will affect individual patients decisions as to whether the risks are acceptable in exchange for probable benefit" --> FDA would consider the consequences of the patient preference studies --> Patient preference specific --> Manufactures (etc.) should submit preference information (PPI) and PPI-related materials to help FDA in decision-making process.


 
Posted : 17/11/2017 5:51 am
(@akashranpura)
Posts: 39
Eminent Member
 

When looking at risk-benefit analysis, one is basically answering the question "is the risk worth it?" To answer this question a thorough analysis needs to be done on the trade-offs of the product. Statistical measures need to be taken in order to fully understand and implement the risks versus reward of the product. When dealing with medical devices the most common risk examined is the chances of the product being fatal or harming the patient to a large degree. However, if the reward is that most people will benefit tremendously from this product, certain executive level decisions need to be made. The magnitude and type of benefits the patient will experience must be examined to a certain degree in this study. Also the length of how effective the product will be is something that should also be taken into consideration.


 
Posted : 17/11/2017 11:06 am
(@hm243)
Posts: 85
Trusted Member
 

As stated previously, the risk-benefit analysis is an important, when determining the benefits and risks that need to be considered for a product. One of the factors that is considered in the analysis is postmarket data. When the product is used in reality, a better assessment can be made of the risks and benefits it poses. With the collection of postmarket data, it will help to identify more benefits and risks for the product and how it affected the patients. FDA has the ability to request post-approval studies. With the data that is collected once the product hits the market will help to improve the benefits and allow for fewer risks in the product. This postmarket data will also take into perspective the response of the patients and how the product affects them. This will allow them to improve the product accordingly to get the best results. Along with the other factors mentioned, risk-benefit analysis can improve and help to create most effective product.


 
Posted : 17/11/2017 7:37 pm
(@alexandrabuga)
Posts: 149
Estimable Member
 

One Factor is Disease chronicity as described in "Factors to Consider When Making Benefit-Risk Determinations in Medical Device Premarket Approval and De Novo Classifications" For example if a patient has a chronic disease but they have "adapted to their illness" and it doesn't effect their daily lives as much they are less likely to tolerate risky devices and is risk averse. Whereas a patient who has a debilitating chronic illness is likely to tolerate a higher risk device. Another factor is the availability of alternate treatments available. So a patient with a life-threatening illness may tolerate more risk for even a small benefit because there aren't alternatives available.


 
Posted : 18/11/2017 12:33 pm
(@srg36)
Posts: 117
Estimable Member
 

Another factor to consider when making benefit-risk determination suggested by FDA in this guidance document is whether the device is a novel technology addressing an unmet medical need. This means that the device has a breakthrough technology by providing a clinically meaningful advantage over existing technologies, posing less risk than an existing treatment, or providing a treatment where no other alternative is available. I think this is an important assessment when it comes to experimental drugs for cancer. The risk may be high, but if there is no other alternative, it may be determined that the potential benefits may outweigh the potential risks.


 
Posted : 18/11/2017 4:47 pm
 ec52
(@ec52)
Posts: 72
Trusted Member
 

To add to other posts on this forum, another factor to consider during benefit-risk analysis is patient perspectives. If the risks are identifiable and definable, risk tolerance will vary among patients, and this will affect individual patient decisions as to whether the risks are acceptable in exchange for a probable benefit. As mentioned in the guidance within the first question on this forum, FDA recognizes that patient perspectives on benefits and risks may reveal reasonable patients who are willing to tolerate a very high level of risk to achieve a probable benefit, especially if that benefit results in an improvement in quality of life. Therefore, manufacturers should take into account patient's willingness and unwillingness to use a device or tolerate risk in exchange for probable benefit, and/or evaluate how patients view trade-offs between benefits and risks of various treatment options (if available).


 
Posted : 18/11/2017 6:34 pm
(@woolynn)
Posts: 36
Eminent Member
 

Risk-benefit analysis is the analysis that seeks to quantify the risk and benefits and hence their ratio.

Analyzing a risk can be heavily dependent on the human factor. A certain level of risk in our lives is accepted as necessary to achieve certain benefits. For example, driving an automobile is a risk most people take daily, also since it is mitigated by the controlling factor of their perception of their individual ability to manage the risk-creating situation. When individuals are exposed to involuntary risk (a risk over which they have no control), they make risk aversion their primary goal. Under these circumstances, individuals require the probability of risk to be as much as one thousand times smaller than for the same situation under their perceived control (a notable example being the common bias in the perception of risk in flying vs. driving)


 
Posted : 19/11/2017 8:05 am
(@puneet)
Posts: 80
Trusted Member
 

Risk-benefit analysis is an important consideration for a product. Part of the risk management framework is to decide what risks that you are willing to take with a product. It is important to identify what kind of risks that you are willing to take and which risks, if any you will take. This is when risk-benefit analysis likely comes into play as companies will have to consider the risk-reward ratio for all factors that they are considering. It is therefore important to identify low-risks that may be worth the reward and high or intermediate risks that don't offer enough of a reward to consider taking such a risk.


 
Posted : 19/11/2017 6:54 pm
(@ppp23)
Posts: 43
Eminent Member
 

The benefit-risk factors may be considered when device manufacturers evaluate appropriate responses to nonconforming product or regulatory compliance issues, such as determining whether to limit the availability of a medical device (e.g., a voluntary recall or market withdrawal). FDA may consider the benefit-risk factors during, for example, evaluation of device shortage situations, selection of the appropriate regulatory engagement mechanism following an inspection during which regulatory non-compliance was observed, evaluation of recalls, and consideration of petitions for variance from those sections of the Quality System (QS) regulation (21 CFR part 820) for which there were inspectional observations during a Premarket Approval (PMA) pre-approval inspection.


 
Posted : 19/11/2017 7:23 pm
(@ao242)
Posts: 43
Eminent Member
 

One of the factors of Risk-benefit analysis according to FDA provided guidance is Patient Preference. It explains the importance of knowing any available patient preference information. Also to consider the modality of treatment/diagnostic procedure and the risk tolerance. FDA would disapprove the IDE application in accordance with 21 CFR 812.30(b), if the potential risks outweigh the anticipated benefits for all subjects.


 
Posted : 19/11/2017 8:43 pm